Re: {OIG} calibration procedures of mass / vortex flow meters

Expand Messages
  • ram prasad
    .Dear all  Flow Verification-        Via Fiedcheck. This is flow simulator, with which sensor and transmitter physical characteristics are checked and
    Message 1 of 3 , Mar 26, 2012
    .Dear all
     Flow Verification-        Via Fiedcheck. This is flow simulator, with which sensor and transmitter physical characteristics are checked and certified.  This can be performed for E+H flowmeters only.

    2. Flow Verification-        Via Ultrasonic Clamp on Flowmeter- Accuracy 5%; This can be performed to check the calibration for third party flowmeters as well
    3. Calibration-        Via Mobile Calibration Rig- This requires the meter to be removed from service and installed on the Rig in series with Master meter and calibrated
    regards
    vram




    --- On Tue, 20/3/12, Manuel Pérez <manuel.perez52@...> wrote:

    From: Manuel Pérez <manuel.perez52@...>
    Subject: Re: {OIG} calibration procedures of mass / vortex flow meters
    To: "ourinstrumentationgroup@yahoogroups.com" <ourinstrumentationgroup@yahoogroups.com>
    Date: Tuesday, 20 March, 2012, 10:18 AM

     
    Hi Anil:

    I don't know if there are such calibration formats, probably yes, but in the case of the Coriolis mass flowmeters you are measuring the weight of the fluid so you must use a scale preferably electronic and run tests weighting batches of the fluid in containers with a known volume, for example and depending the flow you want to measure and the size of the meter, I have used 200 lts. drums and make tests of 150 or 180 lts. depending on the density of the fluid to be measured. Obviously it's only possible in liquids that are no harmful or toxic because the drums are open, if this is the case you will have to be creative and find the way to weight the volumes you want to calibrate.

    In the case of the vortex meters, and depending of what is the fluis you want to measure, again you have to find the way to compare the reading of the meter with known volumes because in this case the vortex meter measures volume. In my experience I have had a lot of troubles with the vortex meters, because when the flowrates change drastically the meters just don't keep the accuracy and fail continuously giving big differences against the expected values.

    I hope this information will help you and if you have further comments, let me know.

    Best regards,

    Ing. Manuel Pérez Schmill
     
    INTECH
    Instrumentación y Tecnología.
    Tel.: (33)1380-6226; Fax: (33)3044-8639.
    Cel.: 3331064-255
    Nextel: (33)1660-8605; ID: 92*13*26414


    De: Anil Inani <inani.anil@...>
    Para: "ourinstrumentationgroup@yahoogroups.com" <ourinstrumentationgroup@yahoogroups.com>
    Enviado: Sábado, 17 de marzo, 2012 3:02:39
    Asunto: {OIG} calibration procedures of mass / vortex flow meters

     
    Hi
     
    Does any one have the calibration procedures of mass / vortex flow meters in industries and sample formats of calibration.

    Regards
    Anil


Your message has been successfully submitted and would be delivered to recipients shortly.