Measuring IVRT

Expand Messages
  • Clifford Thornton
    QUESTION:   What is the best method for measuring the IVRT?   To measure the IVRT, is the CW or PW method more preferable?    Thanks,   Cliff QUESTION:
    Message 1 of 7 , Dec 12, 2008
    • 0 Attachment
      QUESTION:
       
      What is the best method for measuring the IVRT?
       
      To measure the IVRT, is the CW or PW method more preferable? 
       
      Thanks,
       
      Cliff

    • Paula gehr
      According to Gulf Coast Institute  which I just attended last week the answer is CW. If anyone needs a refresher or just beginning ECHO that already has a
      Message 2 of 7 , Dec 19, 2008
      • 0 Attachment
        According to Gulf Coast Institute  which I just attended last week the answer is CW. If anyone needs a refresher or just beginning ECHO that already has a registry or is a MD wanting to learn ECHO I highly recommend this course. Very good lecture and hands on.
                                                                                          Merry Christmas,
                                                                                                  pg

        --- On Fri, 12/12/08, Clifford Thornton <cmt51597@...> wrote:
        From: Clifford Thornton <cmt51597@...>
        Subject: [echocardiography] Measuring IVRT
        To: echocardiography@yahoogroups.com
        Date: Friday, December 12, 2008, 5:00 PM

        QUESTION:
         
        What is the best method for measuring the IVRT?
         
        To measure the IVRT, is the CW or PW method more preferable? 
         
        Thanks,
         
        Cliff


      • rshado
        Either CW or PW may technically be used, however much of the published work used pulsed wave Doppler for myocardial performance for either ventricle. PW is
        Message 3 of 7 , Dec 20, 2008
        • 0 Attachment
          Either CW or PW may technically be used, however much of the
          published work used pulsed wave Doppler for myocardial performance
          for either ventricle.

          PW is more regionally sensitive and usually can be set to obtain
          superior demarcation of valve opening and closure events. Increasing
          sample size may helps, as well as using rejection filters for
          increased noise.

          Most textbooks you will see PW waveforms demonstrating the
          measurement of IVRT/IVCT.

          Rick

          --- In echocardiography@yahoogroups.com, Clifford Thornton
          <cmt51597@...> wrote:
          >
          > QUESTION:
          >  
          > What is the best method for measuring the IVRT?
          >  
          > To measure the IVRT, is the CW or PW method more preferable? 
          >  
          > Thanks,
          >  
          > Cliff
          >
        • TJ
          ... the answer is CW. If anyone needs a refresher or just beginning ECHO that already has a registry or is a MD wanting to learn ECHO I highly recommend this
          Message 4 of 7 , Dec 23, 2008
          • 0 Attachment
            --- In echocardiography@yahoogroups.com, Paula gehr
            <smart_girl1998@...> wrote:
            >
            > According to Gulf Coast Institute  which I just attended last week
            the answer is CW. If anyone needs a refresher or just beginning ECHO
            that already has a registry or is a MD wanting to learn ECHO I highly
            recommend this course. Very good lecture and hands on.
            >
                                                                                              Merry
            Christmas,
            >
                                                                                                     
            pg
            >
            > --- On Fri, 12/12/08, Clifford Thornton <cmt51597@...> wrote:
            >
            > From: Clifford Thornton <cmt51597@...>
            > Subject: [echocardiography] Measuring IVRT
            > To: echocardiography@yahoogroups.com
            > Date: Friday, December 12, 2008, 5:00 PM
            >
            >
            >
            >
            >
            >
            >
            >
            >
            >
            >
            > QUESTION:
            >  
            > What is the best method for measuring the IVRT?
            >  
            > To measure the IVRT, is the CW or PW method more preferable? 
            >  
            > Thanks,
            >  
            > Cliff
            >

            I prefer to measure IVRT and IVCT using a clean tissue Doppler sample
            of the mitral annulus. I find CW to be too cluttered with spurious
            echos even if the gain is optimized. Many machines include these time
            measurements under the tissue Doppler setting.

            PEACE and Merry Christmas

            TJ Whelan RDMS, RDCS, RVT, FASE
          • Mark Harry
            I have found that If the patient is in sinus rhythm, the most accurate way is to use PW Doppler. Position your sample volume at the level of the aortic valve
            Message 5 of 7 , Dec 23, 2008
            • 0 Attachment
              I have found that If the patient is in sinus rhythm, the most accurate way is to use PW Doppler. Position your sample volume at the level of the aortic valve in the apical window and measure the time from the Q wave on the ECG to the closur of the aortic valve. Then place your sample volume at the level of the mitral valve in the apical window and measure from the Q wave on the ECG to the opening of the mitral valve.  The difference between the two measurements is the IVRT.  Always sweep at the fastest sweep speed.  

              Mark Harry RDCS
              Cardiac Ultasound Consulting

              On Dec 12, 2008, at 4:00 PM, Clifford Thornton <cmt51597@...> wrote:

              QUESTION:
               
              What is the best method for measuring the IVRT?
               
              To measure the IVRT, is the CW or PW method more preferable? 
               
              Thanks,
               
              Cliff

            • Udaya kumar
              Thanks Mark for the good information Could you please tell me what the normal IVR time is? When can we assume there is pulmonary hypertension? What s wrong
              Message 6 of 7 , Dec 26, 2008
              • 0 Attachment
                Thanks Mark for the good information
                Could you please tell me what the normal IVR time is?
                When can we assume there is pulmonary hypertension?
                What's wrong with using CW doppler?
                ..Udayakumar


                From: Mark Harry <mharry@...>
                To: "echocardiography@yahoogroups.com" <echocardiography@yahoogroups.com>
                Sent: Wednesday, 24 December, 2008 5:07:12 AM
                Subject: Re: [echocardiography] Measuring IVRT

                I have found that If the patient is in sinus rhythm, the most accurate way is to use PW Doppler. Position your sample volume at the level of the aortic valve in the apical window and measure the time from the Q wave on the ECG to the closur of the aortic valve. Then place your sample volume at the level of the mitral valve in the apical window and measure from the Q wave on the ECG to the opening of the mitral valve.  The difference between the two measurements is the IVRT.  Always sweep at the fastest sweep speed.  

                Mark Harry RDCS
                Cardiac Ultasound Consulting

                On Dec 12, 2008, at 4:00 PM, Clifford Thornton <cmt51597@yahoo. com> wrote:

                QUESTION:
                 
                What is the best method for measuring the IVRT?
                 
                To measure the IVRT, is the CW or PW method more preferable? 
                 
                Thanks,
                 
                Cliff



                Add more friends to your messenger and enjoy! Invite them now.
              • Mark Harry
                Normal is around 70 to 100 milliseconds. Patients with impaired relaxation will have an IVRT over 100 milliseconds. You can use CW Doppler but I believe the
                Message 7 of 7 , Dec 29, 2008
                • 0 Attachment
                  Normal is around 70 to 100 milliseconds.  Patients with impaired relaxation will have an IVRT over 100 milliseconds.  You can use CW Doppler but I believe the temporal resolution of the CW technique is not as good, particularly in a patiens with a dilated cardiomyopathy.

                  Mark J. Harry RDCS
                  Cardiac Ultrasound Consulting


                  On Dec 26, 2008, at 8:46 AM, Udaya kumar wrote:


                  Thanks Mark for the good information
                  Could you please tell me what the normal IVR time is?
                  When can we assume there is pulmonary hypertension?
                  What's wrong with using CW doppler?
                  ..Udayakumar


                  From: Mark Harry <mharry@...>
                  To: "echocardiography@ yahoogroups. com" <echocardiography@ yahoogroups. com>
                  Sent: Wednesday, 24 December, 2008 5:07:12 AM
                  Subject: Re: [echocardiography] Measuring IVRT


                  I have found that If the patient is in sinus rhythm, the most accurate way is to use PW Doppler. Position your sample volume at the level of the aortic valve in the apical window and measure the time from the Q wave on the ECG to the closur of the aortic valve. Then place your sample volume at the level of the mitral valve in the apical window and measure from the Q wave on the ECG to the opening of the mitral valve.  The difference between the two measurements is the IVRT.  Always sweep at the fastest sweep speed.  

                  Mark Harry RDCS
                  Cardiac Ultasound Consulting

                  On Dec 12, 2008, at 4:00 PM, Clifford Thornton <cmt51597@yahoo. com> wrote:


                  QUESTION:
                   
                  What is the best method for measuring the IVRT?
                   
                  To measure the IVRT, is the CW or PW method more preferable? 
                   
                  Thanks,
                   
                  Cliff